Part of: Mechanical Ventilation

  • Mode
  • Fraction of inspired oxygen
  • Tidal volume ()
  • Respiratory rate (RR)
  • Positive End-Expiratory Pressure (PEEP)
  • Pressure Support (PS)
  • Flow shape/contour
  • Inspiratory:Expiratory (I:E) ratio

Control Variables

  • The variable which the ventilator uses as feedback signal for controlling inspiration
  • Flow cannot be a control variable because flow is a volume over time and thus when volume is controlled, flow is controlled indirectly
  • Pressure control maintains a stable pressure in the face of fluctuating respiratory performance, which prevents lung injury from excess pressure but doesn’t give a consistent minute ventilation
    • Advantages
      • Increased mean airway pressure β‡’ improves oxygenation
      • Increased duration of alveolar recruitment β‡’ square pressure waveform causes alveoli to open earlier and remain open for longer allowing for better gas exchange
      • Prevents excessive airway pressures thereby protecting against barotrauma
      • Avoids regional alveolar overdistention
      • May lead to earlier liberation from mechanical ventilation
      • Allows for significant air leak (e.g. during bronchoscopy or bronchopleural fistula)
    • Disadvantages
      • Tidal volume is variable and dependent on respiratory compliance
      • Uncontrolled volume may result in β€œvolutrauma” (overdistension)
      • A high early inspiratory flow may breach the pressure limit if airway resistance is high

  • Volume control gives a more stable minute ventilation, keeping at the desired level
    • Advantages:
      • Guaranteed tidal volumes produces a more stable minute volume
      • The minute volume remains stable over a range of changing pulmonary characteristics
      • The initial flow rate is lower than in pressure-controlled modes, avoiding a high resistance-related early pressure peak
    • Disadvantages:
      • The mean airway pressure is lower with volume control ventilation
      • Recruitment may be poorer in lung units with poor compliance
      • In the presence of a leak, the mean airway pressure may be unstable
      • Insufficient flow may give rise to patient-ventilator dyssynchrony

Targeting Scheme

  • Set point: the ventilator will try to achieve the parameter (control variable chosen)
  • Dual targeting: the ventilator switches from targeting one control variable to another in the middle of the breath
    • For example a breath may start with a pressure control variable using a decelerating flow waveform, then reach the pressure limit mid breath and change to volume control until the target volume is reached
  • Adaptive targeting:
    • For example in PRVC, the inspiratory pressure is automatically adjusted to achieve an average tidal volume target; this varies from breath to breath adapting to the changing compliance (guarantees a prescribed volume while maintaining a square pressure waveform)

Phase Variables

Trigger Variable

  • Determines when a breath is delivered, distinguishing β€˜mandatory’ and β€˜spontaneous’ modes of ventilation
  • Sensitivity affects:
    • Work of breathing
    • Patient-ventilator synchrony

Time Triggered

  • Mandatory ventilation
  • Guarantees a minute volume offering predictable removal and decreased work of breathing
  • Less comfortable and sedation requirements are higher

Flow Triggered

  • Patient effort changes circuit flow
  • Most comfortable but can be over-sensitive leading to dyssynchrony
  • Ventilation is triggered when flow is diverted to the patient when the patient begins inspiration
    • The exact threshold value can be altered but is generally 1-2 L/min1
  • Some machines indicate spontaneous respiratory effort by colouring the waveform

  • Advantages:
    • Generally quite sensitive meaning that patient’s work of breathing is not wasted on triggering the ventilator
    • Allows the patient to have control over their minute volume
    • May decrease the work of breathing
    • More comfortable
    • Permits a lower level of sedation
  • Disadvantages:
    • May be too sensitive, giving rise to auto-triggering (this is probably the only use case for pressure trigger)
    • Does not guarantee a minute volume β€” therefore unsuitable for patients with a diminished or unreliable respiratory drive

Pressure Triggered

  • Patient generates negative pressure
  • A typical pressure trigger threshold would be 1 cm

  • Disadvantages
    • Requires more effort to trigger the ventilator
    • Represents a wasted respiratory effort
    • Less comfortable for the patient
  • Use cases
    • Can be used to decrease auto-triggering (e.g. circuit leak, bronchopleural fistula or hyperdynamic circulation)
    • Can be used to test the power of respiratory musculature in the context of an assessment of readiness for extubation (e.g. a patient who is able to trigger the ventilator by generating a negative intrathoracic pressure of -20 cm is unlikely to fail extubation due to weakness of their respiratory muscles)

Shape-Signal Triggering

  • May decrease wasted effort by β€œpredicting” the next respiratory effort but not widely available

Limit/Target Variable

  • Refers to the maximum value a variable can reach during inspiration but importantly, it does not end inspiration
  • Common limit variables include pressure, flow and volume
    • For example, continue holding this pressure or flow value (don’t exceed it) during the rest of inspiration

Cycling Variable

Time-Cycled

  • Feature of mandatory modes
  • Usually set by setting a respiratory rate and the I:E ratio
  • Advantages:
    • Careful control of minute volume, with obvious advantages for scenarios where tight PaCO2 control is desirable (e.g. traumatic brain injury)
    • Ventilation is unaffected by changes in lung compliance or airway resistance because the timing of the breath is unrelated to any respiratory system parameters
    • Minute ventilation is not affected by an unreliable respiratory drive, making this method suitable for paralysed or deeply unconscious patients
  • Disadvantages:
    • Unsuitable for lightly sedated and awake patients
    • May result in patient-ventilator dyssynchrony particularly if the patient tries to exhale before the cycle time runs out

Flow Cycled

  • The ventilator cycles into the expiratory phase once the flow has decreased to a predetermined value during inspiration
    • Can be expressed as a fixed value in litres per minute or a percentage fraction of the peak flow rate

  • In patients with restrictive lung disease (poor compliance), flow rate drops quickly β†’ lower tidal volumes
  • In patients with emphysema (high compliance), flow rate drops slowly β†’ higher tidal volumes
  • Advantages:
    • More comfortable for the patient
    • Limited by changes in lung compliance and airway resistance, preventing inadvertent ventilator-induced lung injury
  • Disadvantages
    • Tidal volumes may be poor in patients with poor lung compliance, resulting in inadequate minute volume
    • Patient comfort depends on intelligent settings; inappropriately low or high settings could result in uncomfortably deep inspiration or β€œdouble triggering”

Pressure Cycled

  • Largely obsolete in the modern era
  • Advantages: Safety from pressure-related lung injury; decelerating ramp flow pattern; compliance determines cycling
  • Disadvantages: Volume is determined by compliance; respiratory rate may fluctuate; may increase respiratory effort; pressure cannot also be a control variable

Volume Cycled

  • Largely obsolete in the modern era
  • Inspiratory phase ends when the specified volume has been delivered
  • Major disadvantage: propensity to generate high peak airway pressures when lung compliance decreases

Fraction of Inspired Oxygen

  • should be titrated to the lowest value which still maintains adequate oxygenation
  • In common practice, set at 100% after the patient is first intubated, then titrated downward over one to several hours as indicated by pulse oximetry and/or serial ABGs
  • leads to oxygen toxicity; if adequate oxygenation requires , additional strategies are required:
    • ↑ PEEP
    • Recruitment manoeuvres
    • Trial of a different mode

Tidal Volume

  • The tidal volume is 7 mL/kg in a normally breathing patient
  • Most applicable to volume cycled modes (AC, SIMV)
  • Initial values should be weight based:
Patient
Healthy lungs (e.g. neurological catastrophe, drug overdose)10 mL/kg IBW
COPD8 mL/kg IBW
ARDS6 mL/kg IBW
  • Higher leads to ↓ , ↑ pH and ↑ and vice versa

Respiratory Rate

  • Typical respiratory rate is 10-20 breaths/min in order to provide 7-10 L/min of minute ventilation
  • The normal resting minute volume is 70-100 mL/kg/min; therefore to produce tidal volumes of 6-8 mL/kg, a respiratory rate of between 12-16 breaths per minute
  • Higher RR leads to ↓ , ↑ pH and higher risk of auto-PEEP and vice versa

Positive End-Expiratory Pressure

  • Continuous positive pressure present throughout all of ventilation
  • Physiologic effects:
    • ↑ oxygenation
      • ↑ alveolar recruitment
      • ↑ alveolar surface area
    • ↑ cardiac output in CHF patients but can ↓ BP in non-CHF patients
      • ↓ preload
      • ↓ LV afterload
    • ↑ RV afterload β†’ ↑ R-L shunts if present
Figure 5. Relationship between oxygen delivery and PEEP. There exists a sweet spot where an optimal PEEP delivers optimised oxygen delivery
  • In clinical practice: PEEP is set to the lowest value that allows to be ≀ 60% with a minimum value of 5 cm of
  • The healthy lung should be ventilated with 5-8 cm of of PEEP
  • ARDS patients will require higher PEEP (usually > 12 cm of )
  • With bronchospasm, low PEEP or zero PEEP (ZEEP) is often warranted (< 5 cm of )

Pressure Support

  • Amount of additional positive pressure beyond PEEP that is provided during inspiration. Important in PSV, BiPAP and almost always used in SIMV
  • Simple estimate of optimal PS is:
  • However in practice, PS is typically set to twice PEEP

Flow Shape/Contour

  • Describes the pattern of airflow during inspiration; set in volume-targeted ventilator modes
    • Always decelerating shape in pressure-targeted modes as a consequence of lung mechanics
Figure 6. Decelerating flow contour
Figure 7. Constant flow contour
ContourDead space ()Auto-PEEP
Decelerating↓↑↓More
Constant↑↓↑Less

Inspiratory : Expiratory Ratio

  • Ratio between the amount of time spent in inspiration and expiration
  • Normal I:E ratio is 1:2; deviations from this are uncomfortable, often requiring deep sedation
ChangeEffects
↑ Inspiratory time↑ Mean airway pressure β†’ better oxygenation (but ↓ COβ‚‚ clearance, ↑ haemodynamic instability, ↑ gas trapping)
↑ Expiratory time↑ COβ‚‚ clearance β†’ better ventilation (but ↑ probability of atelectasis)
  • In AC and SIMV: usually set indirectly via and flow rate/pattern
  • In PCV: usually set directly
  • In PSV: generally outside of clinician control
  • A higher ratio (higher inspiratory time) results in ↑ and higher risk of auto-PEEP and vice versa

Typical Initial Ventilator Settings

OptionTypical Settings
ModeIntrinsic hyperventilation β†’ SIMV; No intrinsic hyperventilation β†’ AC or SIMV
FiOβ‚‚Start at 100%; Taper as able to 35–60% to keep >60–80 mmHg
Tidal Volume ()~10 cc/kg normal; ~8 cc/kg COPD; ~6 cc/kg ARDS (Use IBW); Lower if > 30 cm
Rate10–20 breaths/min to achieve MV of 7–10 L/min; adjust based on pH
PEEPStart at 5 cm ; Titrate up if <60 on >60% FiOβ‚‚; May start with no PEEP in pure hypoventilation
Pressure Support (n/a for AC)5–20 cm ; Optimal ; (Minimum 5 cm PS always to overcome ETT resistance)

Typical Initial Ventilator Alarm Settings

  • High pressure limit: 10–15 cm Hβ‚‚O above PIP
  • Low pressure limit: 5–10 cm Hβ‚‚O below PIP
  • Low PEEP limit: 3–5 cm Hβ‚‚O below set PEEP
  • Low exhaled tidal volume: 100 mL or 50% below set VT
  • Low minute ventilation: 2–5 L/min or 50% below baseline
  • High minute ventilation: 50% above baseline
  • FiOβ‚‚ alarm: Β±5% from set oxygen concentration
  • Temperature alarm: Β±2Β°C from set temperature
  • Apnea delay: About 20 seconds

Footnotes

  1. Normal mean inspiratory flow rate at rest is around 15 L/min with a peak of around 30-35 L/min ↩